As stated by others, you'll need a stage micrometer that is calibrated with traceability to NIST. You should also review ASTM E1951 and write your own in-house procedure based on ASTM E1951. We are A2LA accredited for testing and we calibrate our own microscope, so we have to comply with the uncertainty reporting requirements as well, which can take some work to get in place if you haven't done that before. If it's not required to be accredited to ISO 17025, then you could skip the uncertainty reporting aspect.
I'd recommend adding some preventive maintenance aspects to your procedure as well, since most audits look at PM records when auditing calibration records of microscopes.
Paul Elliott
Arconic
Eureka MO
Sent: 03-03-2020 11:16
Subject: Metallurgical Microscopes Calibration
Hi everybody,
I'm trying to put together a procedure for microscopes calibration and verification. If we choose to perform the calibration and verification internally by using a certified micrometer, what frequency each of you recommends and if there is any body from an accredited lab or NADCAP certified lab I would like to know what they have in place.
Appreciate your help,
Neda Mandkarian
McLAren Engineering
Livonia